PROBLEM WITH SCANER

#1 User is offline   cretedens 

  • Newbie
  • Pip
  • Group: Members
  • Posts: 2
  • Joined: 30-August 06
  • Gender:Male
  • Location:CRETE CREECE

Posted 07 February 2010 - 10:09 PM

FEW MONTHS AGO THE REPRESENTATIVE OF SIRONA IN GREECE,SENT ME ONE CLEANING SET IN ORDER TO CLEAN THE INTERNAL TUBES OF INLAB. FEW DAYS LATER, WHEN THE SCANNER BEGAN WORGING,ON THE H/Y SCREEN A MESSAGE APPEARED - "THE SCANNER WINDOW IS DIRTY" ALTHOUGHT IT WAS NOT.(THE RED SPOT APPEARS ON THE WINDOW NORMALLY).I ALSO SEE THAT THE SCANNER'S WINDOW COLOUR IS NOT THE SAME AS IT WAS BEFORE {NOT SO DARK}.SIRONA IN GREECE SAYS THAT I MUST CHANGE THE WHOLE LAZER - IS THERE ANY OTHER SOLUTION OR ANY IDEA IN ORDER TO FIND A SOLUTION TO MY PROBLEM?
0

#2 User is offline   Alessandro Devigus 

  • Advanced Member
  • PipPipPip
  • View gallery
  • Group: Admin
  • Posts: 3,394
  • Joined: 20-February 04
  • Gender:Male

Posted 08 February 2010 - 07:47 AM

well that's a good question ... I don't know of any "laser testing" device ...
0

#3 User is offline   joegarciaar 

  • Member
  • PipPip
  • Group: Premium Member
  • Posts: 23
  • Joined: 08-March 04
  • Gender:Male
  • Location:Buenos Aires
  • Interests:The New Dentistry

Posted 08 February 2010 - 02:53 PM

Clean all chambre (only with water and tisuee) In particular the black zone at which point the laser at startup. After calibrating. What is your milling? Good luck



View Postcretedens, on 07 February 2010 - 06:09 PM, said:

FEW MONTHS AGO THE REPRESENTATIVE OF SIRONA IN GREECE,SENT ME ONE CLEANING SET IN ORDER TO CLEAN THE INTERNAL TUBES OF INLAB. FEW DAYS LATER, WHEN THE SCANNER BEGAN WORGING,ON THE H/Y SCREEN A MESSAGE APPEARED - "THE SCANNER WINDOW IS DIRTY" ALTHOUGHT IT WAS NOT.(THE RED SPOT APPEARS ON THE WINDOW NORMALLY).I ALSO SEE THAT THE SCANNER'S WINDOW COLOUR IS NOT THE SAME AS IT WAS BEFORE {NOT SO DARK}.SIRONA IN GREECE SAYS THAT I MUST CHANGE THE WHOLE LAZER - IS THERE ANY OTHER SOLUTION OR ANY IDEA IN ORDER TO FIND A SOLUTION TO MY PROBLEM?

Jose Garcia
Buenos Aires, Argentina
0

#4 User is offline   Martin 

  • Member
  • PipPip
  • Group: Members
  • Posts: 10
  • Joined: 25-March 05
  • Location:Brasil
  • Interests:Cerec inLab

Posted 25 July 2010 - 09:52 PM

The same message appears on my CEREC since I use the “DentaClin” product of Ivoclar Vivadent.
To resolve the problem momentarily (I have to repeat it every week) I clean the laser window with a peace of cotton and little quantity of alcohol, Then I dry with air. Be careful not to break the window laser, its fragile.
0